BOD sample pre-treatment

Because the BOD test is a bioassay, it is critical to maintain optimal conditions for the "bugs" (bacteria and other microorganisms) to remain viable. Temperature, pH, oxygen levels and the presence of disinfectants can all influence the outcome of the BOD test. Before proceeding with the BOD test, the following items must be checked. In many cases pretreatment maybe required

  • Temperature
  • pH
  • presence of residual chlorine or other disinfectants
  • Dissolved oxygen (DO)

Sample pre-treatment - pH

pH extremes kill (or at least severely injure) the "bugs" (microorganisms). Consequently, you must test each sample (not the diluted sample in the BOD bottle) to confirm that the pH is in the appropriate range before proceeding with the BOD test.

  • Test for proper pH ("pH extremes" kill bugs)
    • pH extremes are defined as less than pH 6 or greater than pH 8.5
    • Adjust pH as needed.
    • If the pH of the undiluted sample is less than 6 or greater than 8.5, adjust the pH to 6.5 -7.5 with 1N H2SO4 or 1N NaOH.
    • Do not dilute sample by >0.5% (1.5 ml in a 300 ml bottle). If more acid or base is needed, use more concentrated solutions (i.e., 5N)
    • Adjusted sample pH must be between 6.5 & 7.5.
  • Always seed samples that have been pH-adjusted.

Sample pre-treatment - disinfection

Check wastewater samples for residual chlorine unless it can be demonstrated that the sample was collected prior to where chlorine is added. Check with the wastewater facility operators to determine if the sample was collected prior to disinfection or not. Commercial laboratories should check all treated wastewater for the presence of chlorine and document this.

Chlorine (Remember: chlorine kills bugs!)

  • If any chlorination process is employed:
    1. Test for residual chlorine
    2. If detected, quench the chlorine residual as per Standard Methods and,
    3. SEED the sample(s)

Other disinfection (UV)

  • If ANY disinfection process is employed (UV, chlorine) SEED the sample(s).

Sample pre-treatment — oxygen levels

Water has a limited capacity to hold oxygen. This capacity, or saturation point, is driven by temperature and barometric pressure. Many people incorrectly believe that Standard Methods establishes the maximum dissolved oxygen concentration at 9.0 mg/L. This is not correct. The point of saturation could occur at a much lower or higher oxygen concentration depending upon the temperature and barometric pressure.

If the sample DO is greater than saturation point when the bottles are placed in the incubator, oxygen will physically flow out of solution and appear to be an oxygen demand. The resulting BOD will be falsely high.

The best way to remove samples with super-saturated DO is to warm the samples to 23°C and then shake them vigorously to "strip" the DO from the water.

Check for super-saturation of oxygen (results in high bias)

  • Determine the DO saturation point at your facility at the analysis temperature and barometric pressure.
    • During winter months the water can hold more oxygen was it is colder
    • Can be a problem at facilities where algae are actively growing (e.g., lagoons)
  • Results in high bias (the 'extra' oxygen quickly comes out of solution during incubation, but is calculated as oxygen used during incubation)
  • Reduce excess DO by shaking sample(s) or aerate with filtered compressed air when at 23 °C

Make sure there is enough oxygen!

Always start with an initial DO close to the saturation point for your facility.

If the DO is low, shake or aerate with filtered compressed air to increase the DO concentration. Remember, starting with a higher initial DO will allow you to cover a wide BOD range with each dilution.

Documentation

  • Residual chlorine
    • If sampled upstream of chlorination, have documentation to prove it (e.g., plant schematic)
    • If not using chlorine (e.g., UV), note such on the benchsheet and SEED all samples.
    • Otherwise, you must document that samples were checked for chlorine, how, and results.
  • pH
    • Need documentation someplace that the pH of ANY sample analyzed is within requirements
    • It's acceptable to use short range pH paper for this check
  • Temperature
    • Need documentation someplace that temperature of ANY sample analyzed is within requirements (20 ± 3°C)
    • It's acceptable to use DO meter thermometers to measure sample temperature

DETERMINING APPROPRIATE DILUTIONS

Because the BOD test is a bioassay technique, it is inherently less accurate and precise than analyte-specific tests such as ammonia and total phosphorus. The BOD test is also limited by capacity of water to hold oxygen. Consequently, the analyst must prepare more than one dilution to ensure the BOD depletion criteria are met (e.g., at least 2 mg/L DO consumed and 1 mg/L residual DO during the 5-day test period).

  • Recommend at least two dilutions (preferably 3 or more).
    • The more dilutions used, the easier it is to identify toxicity problems!
    • WWTPs tend to be more familiar with waste (it is theirs, after all!), thus can get away with fewer dilutions.
  • Commercial labs tend to use more dilutions since they receive many samples and may not be as familiar with the samples they receive.
  • Use dilutions which will result in adequate depletion (depletion of at least 2).
  • Need to use dilutions which will not over-deplete (final DO < 1).

BOD VOLUME ESTIMATION CHART

Assumptions: 8.5 mg/L DOI; meets method depletion requirements
Example: Sample BOD is expected to be about 5 to 25 mg/L

Chart which can be used to determine the best dilutions to be used for a particular sample if the approximate BOD is known

One of the first steps in the BOD testing is to determine the volume needed to prepare dilutions for the 5-day incubation period. The above chart is a convenient way to graphically determine the dilutions (sample volumes) needed for the BOD test. The chart was prepared based on the assumption that the initial DO of the dilution water will be 8.5 mg/L and, there must be at least 2 mg/L depletion during the 5-day test period and at least 1 mg/L residue DO remaining.

The idea is to settle on the best dilution which fits the expected BOD range of the sample (1) and then prepare one lower dilution [more sample] (3) and one greater dilution [less sample] (2).

For example, if the expected BOD is between 5 and 25 mg/L, use sample volumes of 5, 10 and 20 mL. The red horizontal bars show the BOD each sample volume will cover. This type of chart can easily be posted at the BOD workstation as an aid to analysts and technicians.

MORE ON DETERMINING APPROPRIATE DILUTIONS

Estimated BOD5 Suggested Sample Volumes
< 5 mg/L 200, 250, 300 mLs
< 10 mg/L 100, 150, 200 mLs
10 - 30 mg/L 25, 50, 100 mLs
30 - 60 mg/L 15, 25, 50 mLs
60 - 90 10, 15, 25 mLs
90 - 150 mg/L 5, 10, 15 mLs
150 - 300 mg/L 3, 5, 10 mLs
300 - 750 mg/L 1, 3, 5 mLs ***
750 - 1500 mg/L 0.5, 1, 3 mLs ***
1500 - 2500 mg/L 0.25, 0.5, 1 mLs ***

The table above is another aid that may be used by analysts and technicians to determine sample dilution. Either the graphic or tabular dilution guide is recommended to provide consistency among analysts.

MAKING INITIAL DILUTIONS (IF YOU NEED TO USE < 3 MLS)

If less than 3 mL of samples volume is used (<1 percent dilution), a preliminary dilution is required prior to pipetting the sampling into the BOD bottle. This is best done by preparing a simple dilution and then pipetting a portion of the diluted sample into the BOD bottle. For example, if sample volumes of 0.5, 1.0 and 2.0 mL of sample are needed, make a 10-fold preliminary dilution. Prepare the 10-fold dilution by pipetting 10 mL of well mixed sample into a 100 mL volumetric flask (or 10 mL graduated cylinder) containing about 50 mL of BOD dilution water. Bring the flask to 100 mL by adding additional BOD dilution water and then mix thoroughly. The below table illustrates the resulting volumes. Remember: the greater the original sample volume used to make an initial dilution, the more representative the dilution will be of the original sample!

MAKING AN INITIAL 10-FOLD DILUTION

  • 10 mLs sample to 100 mLs total volume (with dilution water)
  • 25 mLs sample to 250 mLs total volume (with dilution water)
  • 50 mLs sample to 500 mLs total volume (with dilution water)
  • 100 mLs sample to 1000 mLs total volume (with dilution water)

Make all dilutions with large-bore volumetric pipets and flasks!

mLs of 10X dilution mLs of original sample
5 0.5
10 1.0
20 2.0
25 2.
50 5.0

GENERAL BOD TIPS

The following are items of important consideration when preparing and performing the BOD test.

  • ROTATE BOD bottles!
    • NEVER use the same bottles for blanks and GGA. Rotate bottles so you can prove your cleaning technique is satisfactory.
    • If you always use the sample bottles for your blank, you will not be able to prove the other bottles are clean.
  • Samples must be brought to 20 ± 3 °C before BOD testing.
    • Auditors will expect that the lab temperature should also be maintained between 17-23°C if you expect to prepare samples for BOD testing
  • Record the lab temperature and barometric pressure daily.
    • You must record the barometric pressure and air temperature of your laboratory if the water-saturated air or air-saturated water techniques are used to calibrate your DO probe.
    • You also need both measurements in order to determine whether or not samples are super-saturated and require pre-treatment.

MEASURING OUT SAMPLES

One of the most critical operations, and an often-overlooked process, in the BOD test is measuring out and transferring the sample to the BOD bottle. Failure to use good technique can cause erratic results. BOD results can also exhibit what many call the toxic effect or sliding BOD results.

pipit types
Pipit types

Use only wide-bore, volumetric pipets for BOD testing. These pipets are designed "to contain" (TC) a specific sample volume and are intended to have the sample blown out after the sample is delivered to the BOD bottle. Do not use narrow-bore volumetric or serological pipets for samples. The narrow bore will act as a filter and will prevent particulates from entering the pipet. This will result in poor sub-sampling and low BOD results.

Wide-bore volumetric pipets are available in a variety of volumes from many scientific specialty companies. However, we recognize that there are few volume choices below 10 mL. Consequently, it may be necessary to use wide-bore serological pipets to cover these volumes. Remember, if you must use serological pipets, only use wide-bore types, use them once and refill between pipettings. For example, if you need 3 mL and 6 mL volumes, don't fill a 10 mL serological pipet and sequentially deliver the 3 mL and 6 mL without refilling the pipet. Remember, particles will settle in the pipet during the sample delivery step. There likely be more particles in the second volume delivered so you will likely have poor agreement between dilutions (think: failure of replicates!). If you need 3 mL and 6 mL, refill the pipet between each of these pipettings.

The BOD test is a tough test to do properly. Don't make life difficult for yourself by using poor pipetting or sub-sampling technique.

  • Do not use more than one pipet for a given sample
    • Example: If using 25 mLs, don't use both a 5 mL and a 20 mL pipet... Use ONE 25 mL pipet
  • DON'T fill a pipet twice to obtain a certain volume
    • Example: If using 200 mLs, don't pipet twice with a 100 mL pipet.
    • Use a graduated cylinder.
    • Remember, error is additive; each time you measure use a pipet or graduated cylinder, there is error. If you pipet twice, you error is double!
    • Keep it simple— only fill your pipet or graduated cylinder once!
  • Serological pipets
    • Ensure that they are wide-bore, and
    • Use them once. Re-fill after each pipetting.
  • Graduated cylinders (> 100 mL)
    • DON'T agonize over "getting exactly to the mark". Use the "rapid attack" approach:
      • Pour quickly
      • Get as close to target volume as you can;
      • Record actual volume used

The longer you agonize over the volume, the more opportunity the particulates in the sample have to settle out. This can bias your sample which can result in either a higher or lower BOD results.

Always re-mix the sample between pipetting out each dilution. If you mix only initially, particles will settle out the longer the sample bottle sits. If you pipet slowly, solids are settling all the while. Each subsequent pipetting will contain fewer solids if pipetting from the top, or more solids if pipetting from the bottom of the bottle.

LARGE SAMPLE VOLUMES REQUIRE EXTRA NUTRIENTS

When larger sample volumes are used to prepare BOD dilutions (>200 mL), it will be necessary to add extra nutrients. Use the following guidance to determine the amount of supplemental nutrients to add:

CURRENT GUIDANCE: Standard Methods
When a bottle contains more than 67% of the sample (> 200 mL) after dilution, nutrients may be limited and subsequently reduce biological activity. The resulting BOD will be biased low. In such samples, add the nutrient/buffer solutions (3a through 3e) directly to each BOD bottle at a rate of 1 mL/L (0.33 mL/300-mL bottle) or use commercially prepared solutions/pillows designed to dose the appropriate bottle size.

When individual nutrient pillows are used, it's OK to use dilution water that may contain some nutrients already.

Make life easier— use the nutrient pillows designed for individual 300 mL BOD bottle.

MEASURING INITIAL DO

Below are important considerations when making the initial DO measurements for the BOD test.

  • It's a good idea to warm up the meter and calibrate first.
  • If the meter has already been used earlier, re-check the calibration.
  • Don't let samples sit too long between adding dilution water and the initial DO measurements.
  • Standard Methods requires measuring the initial DO no longer than 30 minutes after adding the dilution water.
  • Impact of a long delay on samples with rapid or instantaneous demand—
    • you will lose that instantaneous measure.
    • if assess user fees, instantaneous BOD can result in lower fees.
  • Must actually measure the DO of each dilution (vs. measuring initial sample DO and reporting for each dilution).

INCUBATING SAMPLES

The incubation step is a critical part of the BOD test. The 5-day incubation period was originally established by early BOD pioneers based on the length of time it took water to flow from London, England, down the Thames River to the North Sea. Although this may seem like an arbitrary time, it has been internationally adopted as the "standard" incubation period for the BOD test. The following are critical considerations related to the incubation period.

  • Incubate for 5-days (hence the term BOD5) ± 6 hours. Anything beyond that and your results may be questioned.
  • At 20 ± 1 °C (In the dark). Document temperature each day samples are in progress.
  • Fill water seals with dilution water; cap to reduce evaporation.
  • Check daily, add water to seals if necessary.
  • Before removing stoppers, pour off the water in the seals.
SIMPLE LIST OF BOD SAMPLE IN/OUT DAYS.
Day in) Read out on
Monday Saturday
Tuesday Sunday
Wednesday Monday
Thursday Tuesday
Friday Wednesday
Saturday Thursday
Sunday Friday

Due to the 5-day testing period, certain samples require that set-ups and read-outs of results be performed by different individuals.

CALCULATING BOD

The final step in the test is to perform the needed calculations and to evaluate the final BOD results.

updated oxygen saturation table considering both pressure and temperature

Care must be taken to consider all of the below items:

  • Dilutions meet depletion criteria?
    • Residual DO at least 1 mg/L
    • DO depletion at least 2 mg/L
  • Average dilutions meeting depletion criteria.
  • Check for sample toxicity
  • DOi = initial DO
  • DOf = final DO
  • SCF - seed correction factor in mg/L (oxygen contribution from the seed)
  • DF = dilution factor = 300 mL/sample volume used and any other dilution factors